The SPRTCL provides an extensive list of calibration services that cover the whole temperature range for which long-stem SPRTs, including high-temperature SPRTs, are the ITS-90 interpolation artifacts. This range spans from the triple point of argon (-189.3442 °C) to the freezing point of silver (961.78 °C). The laboratory also provides calibrations of capsule-type SPRTs in a limited temperature range, from the triple point of argon up to the freezing point of tin (231.928 °C). The SPRTs are calibrated by employing fixed-point cells, artifacts that realize the ITS-90 by enclosing high-purity materials (99.9999% or more) for which the phase transition temperatures (melting, freezing or triple points) are known with extreme accuracy. Such temperatures are the defining ITS-90 fixed points, and SPRTs are calibrated at combinations of these fixed points that correspond to the desired ITS-90 subrange. The fixed points realized by the SPRTCL are:
| Fixed point | T, K | T, °C | Uncertainty, mK (k=2) | | Triple point of argon | 83.8058 | -189.3442 | 0.14 | | Triple point of mercury | 234.3156 | -38.8344 | 0.15 | | Triple point of water | 273.16 | 0.01 | 0.09 | | Melting point of gallium | 302.9146 | 29.7646 | 0.13 | | Freezing point of indium | 429.7485 | 156.5985 | 0.25 | | Freezing point of tin | 505.078 | 231.928 | 0.27 | | Freezing point of zinc | 692.677 | 419.527 | 0.55 | | Freezing point of aluminum | 933.473 | 660.323 | TBD | | Freezing point of silver | 1234.93 | 961.78 | TBD |
|